Technical data XL3

All specifications comply with the IEC61672 standard. Further standards – as far as they go beyond this standard – are listed with the respective items.

Sound level measurement  

Calibratable product configurations class 1

  • XL3 and the M2340 / M2230 measurement microphone builds an integrating sound level meter with type approval class 1 according to IEC 61672 and ANSI S1.4.

Product configurations class 1

  • XL3 with M2340 / M2230 measuring microphone class 1 according to IEC 61672 and ANSI S1.4;

  • XL3 with M2211 / M2215 measurement microphone class 1 frequency response according to IEC 61672 and ANSI S1.4.

The specifications given apply to operation with the microphone attached or detached.

Product configurations class 2

  • XL3 with M4261 measurement microphone class 2 according to IEC 61672 and ANSI S1.4.

Standards

  • IEC 61672:2014, IEC 61672:2003, IEC 61260:2014, IEC 61260:2003, IEC 60651, IEC 60804;

  • China: GB/T 3785:2010, GB/T 3241, GB 3096-2008, GB 50526, GB-T 4959;

  • Germany: DIN 15905-5, DIN 45657:2014, DIN 45657:2005, DIN 45645-2, optional: DIN 45645-1;

  • Japan: JIS C1509-1:2005, JIS C 1513 class 1, JIS C 1514 class 0;

  • Switzerland: V-NISSG, NAO;

  • UK: BS 4142:2014, BS 5969, BS 6698;

  • USA: ANSI S1.4-2014, ANSI S1.43, ANSI S1.11-2014;

  • International IEC standards have been adapted as European standards and the letters IEC have been replaced by EN. XL3 is compliant with these EN standards.

Weighting

Level details

  • Measurement bandwidth (–3 dB): 4.4 Hz – 23.0 kHz;

  • Level resolution: 0.1 dB;

  • Intrinsic noise: 2.1 μV(Z).

Measuring range with different microphones

  • XL3 + M2340: 17.4 dB(A) – 138.3 dB @ 42 mV/Pa;

  • XL3 + M2230: 17.1 dB(A) – 137.8 dB @ 42 mV/Pa;

  • XL3 + M2215: 25 dB(A) – 153 dB @ 8 mV/Pa;

  • XL3 + M2211: 21 dB(A) – 144 dB @ 20 mV/Pa;

  • XL3 + M2914: 6.5 dB(A) – 103 dB @ 320 mV/Pa;

  • XL3 + M4261: 27 dB(A) – 146 dB @ 16 mV/Pa.

Linear measuring range according to IEC 61672 / ANSI S1.4

  • XL3 + M2340: 25 dB(A) – 138 dB | 28 dB(C) – 138 dB @ 42 mV/Pa;

  • XL3 + M2230: 24 dB(A) – 137 dB | 27 dB(C) – 137 dB @ 42 mV/Pa;

  • XL3 + M2215: 33 dB(A) – 153 dB @ 8 mV/Pa;

  • XL3 + M2211: 29 dB(A) – 144 dB @ 20 mV/Pa;

  • XL3 + M2914: 14 dB(A) – 103 dB @ 320 mV/Pa;

  • XL3 + M4261: 33 dB(A) – 146 dB @ 16 mV/Pa.

Stabilization time

< 10 s.

Integration times

  • Minimum: 1 second (default) or 100 ms (with Extended Noise Measurement option);

  • Maximum: 24 hours.

Intrinsic noise typical without measuring microphone
@ S = 42 mV/Pa

  • Frequency weighting A: 5.1 dBA;

  • Frequency weighting C: 4.1 dBC;

  • Frequency weighting Z: 8.0 dBZ.

Standard functions

  • SPL actual, Leq, Lmin, Lmax, Lpeak, LE;

  • Time weighting Fast, Slow;

  • Broadband, 1/1 Octave and 1/3rd Octave spectral view;

  • Gliding LAeq and LCeq with selectable time window from 1 second to 1 hour;

  • TaktMax according to DIN 45645-1;

  • All measurement results are simultaneously available;

  • Logging of all or data or subsets in selectable intervals ≥ 1 second;

  • Wizard for measuring the correction values for live events of the levels LAeq, LCeq and LCpeak;

  • Individual limit values for each sound level displayed;

  • Recording of compressed audio;

  • Digital I/O interface for controlling accessories.

Functions of Extended Noise Measurement option
  • Time weighting Impulse;

  • Level difference LAIeq – LAeq;

  • Sound exposure level LAE;

  • Time-graph view;

  • Percentiles / levels of the level frequency distribution for broadband and spectral measurements
    Flexible setting from 0.1% to 99.9% with 7 values in parallel
    Sampling rate for Fast/Slow weighted values: every 1.3 ms
    Wideband: with 0.1 dB class bandwidth, based on Lxy sampling (x = A, C or Z, y = F, S or EQ1”)
    1/1 Octave band and 1/3rd Octave band spectrum: in 1.0 dB class width, based on Lxy (x = A, C or Z / y = F or S);

  • 100 ms logging of all or data or subsets;

  • Recording of uncompressed audio.

Spectrum

  • Compliant with class 1 of IEC 61260:2014 and ANSI S1.11-2014 (filter base 10);

  • Octave band display: 8 Hz – 16 kHz;

  • 1/3rd Octave band display: 6.3 Hz – 20 kHz;

  • Selectable frequency range is displayed together with A/Z wideband level;

  • Logging of Leq, min, max every 100 ms or 1 s.

Reverberation Time

 

 

  • Conforms with ISO 3382 and ASTM E2235 based on Schroeder’s backwards integration;

  • Octave bands results from 63 Hz - 8 kHz;

  • Measurement parameters: T20, T30;

  • Impulse and gated noise source;

  • Automatic averaging for each position;

  • Chart and table representation of results;

  • Fixed minimum trigger level: 80 dB LAPK;

  • Warning indicators according to ISO 3382;

  • Range: 10 ms - 60 seconds;

  • Minimum reverberation time (typical):

    • < 100 Hz: 0.3 second;

    • 100 - 200 Hz: 0.2 second;

    • > 200 Hz: 0.1 second.

With the “Extended Room Acoustics” Option

  • 1/3 octave band: 50 Hz - 10 kHz;

  • T20, T30, T15, EDT simultaneously;

  • Calculating spatial room average (Measurement Series) up to 99 positions;

  • Audio recording (32-bit float);

  • Adjustable minimum trigger level from 50 to 100 dB LAPK.

Sound Insulation

 

With the “Sound Insulation” Option

 

Determination of airborne, impact and facade sound insulation on the instrument.

  • Automated data averaging;

  • Results as chart and table.

Airborne Sound Insulation:

  • Sound Sources: Speaker.

  • Standards:

    • ISO16283-1:2014;

    • ISO 717-1: 2021;

    • England/Wales: Approved Document E (2003).

  • Results:

    • Dw | Dn,w | DnT,w | R'w;

    • Spectrum adaption terms C, Ctr;

Impact Sound Insulation:

  • Sound Sources: Tapping Machine, Rubber Ball;

  • Standards:

    • ISO16283-2:2018;

    • ISO 717-2:2020;

    • England/Wales: Approved Document E (2003).

  • Results:

    • With Tapping Machine: L'n,w | L'nT,w;

    • With Rubber Ball: L'IA,Fmax | L'iA, Fmax,V,T;

    • Spectrum adaption terms CI.

Facade Sound Insulation:

  • Sound Sources: Element Loudspeaker, Global Loudspeaker;

  • Standards:

    • ISO16283-3:2016;

    • ISO 717-1: 2021.

  • Results:

    • With Element Loudspeaker: Dw | R'45°,w;

    • With Global Loudspeaker: Dls,2m,w | Dls,2m,n,w | Dls,2m,nT,w;

    • Spectrum adaption terms C, Ctr––.

Calibration

 

Free-field correction

  • NTi Audio class 1 sound calibrator: M2340 / M2230 / M2215 / M2211: –0.10 dB;

  • NTi Audio class 1 sound calibrator with 1/4" calibrator adapter, type: ADP 1/4-P:
    M4260: +0.10 dB;
    M4261: +0.20 dB.

Windscreen correction

@ 1 kHz

  • 50 mm windscreen: +0.03 dB;

  • 90 mm windscreen: –0.04 dB;

  • 150 mm windscreen: –0.04 dB;

  • WP30: –0.03 dB.

Calibration

  • Recommended calibration interval: 1 year;

  • Microphone calibration with external sound calibrator possible;

  • Calibration certificate for a new sound level meter is optionally available.

Input / output interfaces

Audio input

  • XLR balanced:

    • Input impedance 200 kΩ;

    • Phantom power: +48 V switchable; with maximum output current of 10 mA according to IEC 61938;

    • Automatic Sensor Detection (ASD) for NTi Audio measuring microphones and preamplifier MA230 / MA220;

  • Internal speech microphone for recording voice memos.

Audio output

  • Built-in speaker;

  • Headphone socket 3.5 mm stereo; output reference: @ SPL Level 114.0 dBSPL (calibrated microphone) = –12 dBu.

USB-A interface

USB Host supporting the devices described below.

USB-C interface

USB Device supporting MTP (file access from the PC) and Network (website access from the PC), as well as charging the Li-Ion battery.

USB devices

Supported devices:

  • USB-C to LAN adapter, NTi # 600 000 535;

  • 4G/LTE gateways with RNDIS protocol;

  • Mass storage like USB stick, SSD;

  • Vaisala or LCJ Capteurs weather station (see below).

Memory

32 GB micro-SDHC card (default), replaceable, for storing measurement data in ASCII format, as well as audio data (WAV) and screenshots (PNG)

Supported formats: FAT32 and NTFS

Power supply

  • Rechargeable Li-Ion battery:

    • Typ. 3.6 V / 6'000 mAh;

    • Voltage range: 3.0 – 4.07 VDC (theXL3 limits the charging voltage to 4.05 VDC, and thus doubles the number of possible charging cycles);

    • Energy density = 339 Wh/l;

    • Typical battery life @ 25 °C (77 °F) with microphone M2340:
      with display active: >8 h;
      with display switched off: >12 h.

    • Operating temperature: –20 to +60 °C (–4 to +140 °F);

    • The XL3 switches OFF automatically as soon as either the battery charge level drops to 0%, or the temperature of the battery drops below –19 °C (–2.2 °F) or rises above +60 °C (+140 °F). Before an automatic self-shutdown, the XL3 stops the current measurement and saves the present results.

  • Linear external power supply 9 VDC / 2 A:

    • Range: 7.0 – 17.0 VDC @ minimum 4 W;

    • Charges Li-Ion battery in operation; charging time from 10% to 80%: typ. 140 min;

    • Maximum charging power 15 W.

  • USB-C supply with 5 VDC / 1.5 – 3 A / 5 W or 15 W according to USB-C specification release 1.2 is sufficient to operate the XL3 + charge the battery; USB BC1.2 is not supported;

  • USB-A supply with 5 VDC / 0.5 A (e.g via a USB-A to USB-C adapter) does not provide sufficient power to supply the XL3.

Automatic restart

The XL3 automatically turns back ON and resumes the last active measurement:

  1. after an automatic self-shutdown (due to too low charge level), or;

  2. after unintentional removal of the battery (while the device was running).

as soon as it is is reconnected to a voltage source (e.g. power supply unit or charged battery).

Weather station

 
Vaisala
  • WXT532;

  • WXT533;

  • WXT536.

LCJ Capteurs LCJ-CV7 SA-SDI_T.

 

General  

Clock

Real-time clock with own lithium battery; drift < 1.7 s per 24 hrs.

Mechanics

  • 1/4" tripod connection and fold-out stand on rear side;

  • Display: 480 x 800 pixels, 4.3" IPS;

  • Entry: 8 buttons, capacitive multitouch-display;

  • Dimensions L x W x H: 210 x 85 x 45 mm (8.3 x 3.4 x 1.8 ");

  • Weight: 500 g (1.1 lb) including Li-Ion battery.

Temperature

–10 to +50 °C (+14 to +122 °F).

Humidity

5 to 90% RH, non-condensing.

Sensitivity to high frequency fields

Classification group X.

Electromagnetic compatibility

CE according to: EN 61326-1 class B, EN 55011 class B, EN 61000-4-2 to -6 and -11.

Protection class

IP51.

ATEX

  • For applications in Zone 2 hazardous areas according to IEC 60079;

  • Compliant with 2014/34/EU.